An important part of performing instrument calibration is determining the extent of an instrument’s error. Error is usually measured in percent of span.
Calculate the percent of span error for each of the following examples, and be sure to note the sign of the error (positive or negative):
Pressure gauge
• LRV = 0 PSI
• URV = 100 PSI
• Test pressure = 65 PSI
• Instrument indication = 67 PSI
• Error = % of span
Weigh scale
• LRV = 0 pounds
• URV = 40,000 pounds
• Test weight = 10,000 pounds
• Instrument indication = 9,995 pounds
• Error = % of span
Thermometer
• LRV = -40oF
• URV = 250oF
• Test temperature = 70oF
• Instrument indication = 68oF
• Error = % of span
pH analyzer
• LRV = 4 pH
• URV = 10 pH
• Test buffer solution = 7.04 pH
• Instrument indication = 7.13 pH
• Error = % of span
Also, show the math you used to calculate each of the error percentages.
Show Answer
Pressure gauge
• LRV = 0 PSI • URV = 100 PSI • Test pressure = 65 PSI • Instrument indication = 67 PSI
• Error = +2 % of span
Weigh scale
• LRV = 0 pounds • URV = 40,000 pounds • Test weight = 10,000 pounds • Instrument indication = 9,995 pounds
• Error = -0.0125 % of span
Thermometer
• LRV = -40oF • URV = 250oF • Test temperature = 70oF • Instrument indication = 68oF
• Error = -0.69 % of span
pH analyzer
• LRV = 4 pH • URV = 10 pH • Test buffer solution = 7.04 pH • Instrument indication = 7.13 pH
• Error = +1.5 % of span