Thermocouple Sensor Calibration

Thermocouples measure temperature and are used quite often in process controls.

How They Work

When wires of two different thermoelectrically homogeneous materials are joined at one end and placed in a temperature gradient, a thermoelectric voltage (EMF) is observed at the other end. The connection is called the measuring junction. On all thermocouples, the red lead is negative. The color of the other wire indicates the thermocouple type. For example, on a J-type thermocouple, the positive wire is white. Tables for each type of thermocouple list the voltages produced at various temperatures. Thermocouples should be checked whenever there are indications that the output is not accurate. It may also be necessary to check a thermocouple that will be used for a measurement standard.

Input and Output Measurement Standards

A temperature bath provides controlled temperatures for testing a sensor. A well in the temperature bath is used to hold the sensor during the accuracy check. Another well is used to hold a measurement standard thermometer, it is used to confirm the actual bath temperature. A second measurement standard thermometer is used to read the ambient temperature at the reference junction. The sensor output signal and ranges determine the thermocouple output measurement standard. Since the output is measured in millivolts, a millivolt meter is used to read the output.

Connections

Set up the temperature bath as a temperature input standard to the thermocouple. Select the output standard with the appropriate range for reading millivolts. Connect the red lead to the negative millivolt meter input and the white lead to the positive millivolt meter input.

Three Point Checks

Because no adjustments are possible, we can only check the calibration of a thermocouple sensor. This check is generally done at three test point input values: ambient temperature, mid-range temperature, and upper value of the application range. Recall that in a thermocouple, it is the difference in the temperature between the reference and the measuring junction that produces a millvoltage output.

Before inserting the thermocouple into the bath, determine the ambient temperature, which represents the temperature at the measuring junction of the thermocouple. When using look-up tables that are referenced to 0 Deg., you must compensate for ambient temperature. The millivolt value in the tables for the ambient temperature is added to the value from the sensor. This compensated millivolt value is used to determine from the tables the correct temperature.